hexagon logo

Auto calibrate

When you use auto calibrate can you choose the probe angles you want to calibrate? I have a probe with like 20 articulations on file. Job i am running only needs 3. Machine does not have probe changer so i am manualy changing probe to measure small holes. I put auto cali in program so that i can hit enter and walk away to work on other machine. I assume that you can set what probes to auto cal in that unit ? Any help would be appreciated. Thanks
Parents
  • 3) Setup Clearance is a general setting that works on ALL probes, not just the 1 you set it up in.
    Now if you mean, prehit/retract and move/touch speed then yes, those are captured and stored in the parameter sets individually.

    4) The only time you should answer "YES" is to locate with the Master_Probe. Everything else would be "NO" in order to tie the tips all together.
Reply
  • 3) Setup Clearance is a general setting that works on ALL probes, not just the 1 you set it up in.
    Now if you mean, prehit/retract and move/touch speed then yes, those are captured and stored in the parameter sets individually.

    4) The only time you should answer "YES" is to locate with the Master_Probe. Everything else would be "NO" in order to tie the tips all together.
Children
No Data