So, I tried to get my feet a little bit wet with looping, and I'm missing something or doing something wrong here and I can't figure it out. Basically, I've been running one part with a manual alignment and then 2 DCC alignments. That way I could open up the part number, do a manual alignment on the table where the part is, and after the first setup, I can just start the program where DCC begins so I can be more productive by just slapping a new part up there and running it DCC. I want to be able to run more than one part at once by just offsetting it in Y. I have 2 parts on the table, the original part location, and then after the first part is done probing, I ask it to move down 11.5" in Y and re-measure the second part. I have 2 probes that I'm using. It measure with a 4mmx20mm probe first, then changes to a 1mmx40 mm, calibrates, inspects some of the finer holes, then it switches back to the 4mmx20mm probe so the next part can be ready and it calibrates that probe as well. I'm running into a problem where basically It inspects the first part just fine with both probes on all the holes...it then inspects the second part with the 4mmx20mm probe just fine...but when it does a probe change to the 1mmx40mm probe, it tries to calibrate off my sphere...but the sphere has "moved" -11.5 inches in Y- , the same amount of space that the offsets for the loop is telling my program. I can't figure this out, because the 4mmx20 mm probe does just fine on the second part...it's the second probe for some reason that has trouble. I'm going crazy here! I'll post the code so y'all can see it. I will take any suggestions or hints. I know some people might not like how I'm running it? Basically it might be more efficient to run all of the first probe on BOTH parts, then change tools and probe both parts as well. Yes, the tool changing might be excessive. But...it shouldn't affect the program and do what it's doing to me??? This doesn't make sense.

MODE/DCC
V1 =LOOP/START,ID=YES,NUMBER=2,START=1,SKIP=,
OFFSET:XAXIS=0,YAXIS=-11.5,ZAXIS=0,ANGLE=0
MOVESPEED/ 250
TOUCHSPEED/ 3
FLY/ON
MOVE/POINT,NORMAL,<-0.0075,2.3808,7.921>
MOVE/POINT,NORMAL,<4.134,2.3808,2.5197>
PLN2 =FEAT/PLANE,CARTESIAN,TRIANGLE
THEO/<5.7484,1.842,0>,<0,0,1>
ACTL/<5.7485,1.8418,-0.0045>,<0.0002279,-0.0007598,0.9999997>
MEAS/PLANE,3
HIT/BASIC,NORMAL,<3.9052,1.3284,0>,<0,0,1>,<3.9051,1.3 281,-0.0045>,USE THEO=YES
                  (CODE REMOVED TO SAFE SPACE FOR FORUM)
ENDMEAS/
MOVE/POINT,NORMAL,<7.3177,-0.4021,1.109>
MOVE/POINT,NORMAL,<7.3177,-0.4457,-0.5434>
LIN2 =FEAT/LINE,CARTESIAN,UNBOUNDED
THEO/<0.1081,0,-0.6337>,<1,0,0>
ACTL/<0.1079,0.0084,-0.6336>,<1,0.000076,0>
MEAS/LINE,2,ZPLUS
HIT/BASIC,NORMAL,<0.1081,0,-0.6384>,<0,-1,0>,<0.1079,0.0084,-0.6385>,USE THEO=YES
                    (CODE REMOVED TO SAFE SPACE FOR FORUM)
ENDMEAS/
MOVE/POINT,NORMAL,<8.8349,-0.1429,3.0301>
MOVE/POINT,NORMAL,<8.8349,2.464,3.1052>
MOVE/POINT,NORMAL,<-1.0149,2.464,2.9871>
MOVE/POINT,NORMAL,<-0.821,2.464,-0.395>
PNT2 =FEAT/POINT,CARTESIAN
THEO/<0,2.0871,-0.3681>,<-1,0,0>
ACTL/<0.0111,2.0867,-0.368>,<-1,0,0>
MEAS/POINT,1,WORKPLANE
HIT/BASIC,NORMAL,<0,2.0871,-0.3681>,<-1,0,0>,<0.0111,2.0867,-0.368>,USE THEO=YES
ENDMEAS/
A2 =ALIGNMENT/START,RECALL:STARTUP,LIST=YES
ALIGNMENT/LEVEL,ZPLUS,PLN2
ALIGNMENT/TRANS,ZAXIS,PLN2
ALIGNMENT/ROTATE,XPLUS,TO,LIN2,ABOUT,ZPLUS
ALIGNMENT/TRANS,YAXIS,LIN2
ALIGNMENT/TRANS,XAXIS,PNT2
ALIGNMENT/END
TOUCHSPEED/ 1
PREHIT/0.05
RETRACT/0.05
MOVE/POINT,NORMAL,<-0.9451,2.0871,2.4574>
                     (CODE REMOVED TO SAFE SPACE FOR FORUM)
PLN3-DATUM B =FEAT/PLANE,CARTESIAN,TRIANGLE
THEO/<4.7898,11.5,-0.7307>,<0,-1,0>
ACTL/<4.7909,11.5004,-0.7393>,<0.0000402,-0.999999,-0.0014355>
MEAS/PLANE,8
HIT/BASIC,NORMAL,<0.1503,11.5,-0.6915>,<0,-1,0>,<0.1515,11.5001,-0.7002>,USE THEO=YES
                     (PART OF CODE TAKEN OUT TO SAFE SPACE FOR THE FORUM, WAS JUST MEASURING A PLANE)
ENDMEAS/
MOVE/POINT,NORMAL,<8.0074,-0.1971,2.386>
MOVE/POINT,NORMAL,<8.0074,2.5067,2.3753>
PLN4-DATUM A =FEAT/PLANE,CARTESIAN,TRIANGLE
THEO/<5.7571,13.8907,0>,<0,0,1>
ACTL/<5.758,13.8904,-0.0083>,<0.0000803,-0.0002732,1>
MEAS/PLANE,12
               (PART OF CODE TAKEN OUT TO SAFE SPACE FOR THE FORUM, WAS JUST MEASURING A PLANE)
ENDMEAS/
MOVE/POINT,NORMAL,<7.3516,1.0049,3.4112>
              (CODE REMOVED TO SAFE SPACE FOR FORUM)
PLN5-BACKSIDE FOR COMPARISON =FEAT/PLANE,CARTESIAN,TRIANGLE
THEO/<6.2292,16.09,-0.7728>,<0,1,0>
ACTL/<6.2299,16.095,-0.7814>,<-0.0000332,0.9999996,-0.0008982>
MEAS/PLANE,9
HIT/BASIC,NORMAL,<11.4342,16.09,-0.6407>,<0,1,0>,<11.4354,16.0952,-0.6494>,USE THEO=YES
                 (PART OF CODE TAKEN OUT TO SAFE SPACE FOR THE FORUM, WAS JUST MEASURING A PLANE)
ENDMEAS/
MOVE/POINT,NORMAL,<-0.962,5.6591,2.1858>
MOVE/POINT,NORMAL,<-0.5673,5.6591,-0.5041>
PLN3-DATUM C =FEAT/PLANE,CARTESIAN,TRIANGLE
THEO/<0,14.0367,-0.4754>,<-1,0,0>
ACTL/<0.0007,14.0363,-0.4842>,<-0.9999982,0.0000008,0.001918>
MEAS/PLANE,17
HIT/BASIC,NORMAL,<0,16.0305,-0.7623>,<-1,0,0>,<-0.0001,16.0302,-0.7713>,USE THEO=YES
                       (PART OF CODE TAKEN OUT TO SAFE SPACE FOR THE FORUM, WAS JUST MEASURING A PLANE)
ENDMEAS/
MOVE/POINT,NORMAL,<-1.094,0.0399,2.1094>
MOVE/POINT,NORMAL,<2.0704,0.0399,3.2446>
A3 =ALIGNMENT/START,RECALL:STARTUP,LIST=YES
ALIGNMENT/LEVEL,ZPLUS,PLN4-DATUM A
ALIGNMENT/TRANS,ZAXIS,PLN4-DATUM A
ALIGNMENT/ROTATE,YMINUS,TO,PLN3-DATUM B,ABOUT,ZPLUS
ALIGNMENT/TRANS,YAXIS,PLN3-DATUM B
ALIGNMENT/TRANS,XAXIS,PLN3-DATUM C
ALIGNMENT/END
CIR1-LEFT BORE OD =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,OUT,LEAST_SQR
THEO/<2.27,13.795,0.4>,<0,0,1>,2.94,0
ACTL/<2.2748,13.7989,0.3881>,<0,0,1>,2.9428,0
TARG/<2.27,13.795,0.4>,<0,0,1>
START ANG=0,END ANG=360
ANGLE VEC=<1,0,0>
DIRECTION=CCW
SHOW FEATURE PARAMETERS=NO
SHOW CONTACT PARAMETERS=YES
NUMHITS=10,DEPTH=-0.05,PITCH=0
SAMPLE METHOD=SAMPLE_HITS
SAMPLE HITS=0,SPACER=0
AVOIDANCE MOVE=BOTH,DISTANCE=1
FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
SHOW HITS=NO
CIR2-RIGHT BORE OD =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,OUT,LEAST_SQR
THEO/<9.235,13.795,0.4>,<0,0,1>,2.94,0
ACTL/<9.2371,13.7989,0.3881>,<0,0,1>,2.9423,0
TARG/<9.235,13.795,0.4>,<0,0,1>
START ANG=0,END ANG=360
ANGLE VEC=<1,0,0>
DIRECTION=CCW
SHOW FEATURE PARAMETERS=NO
SHOW CONTACT PARAMETERS=YES
NUMHITS=10,DEPTH=-0.05,PITCH=0
SAMPLE METHOD=SAMPLE_HITS
SAMPLE HITS=0,SPACER=0
AVOIDANCE MOVE=BOTH,DISTANCE=1
FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
SHOW HITS=NO
               (CODE REMOVED TO SAFE SPACE FOR FORUM)-CODE FOR MEASURING A BORE 
CIR17-BOTTOM RIGHT CLEARANCE HOLE =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
THEO/<6.276,12.355,-0.315>,<0,0,1>,0.56
ACTL/<6.2787,12.3591,-0.3269>,<0,0,1>,0.566
TARG/<6.276,12.355,-0.315>,<0,0,1>
START ANG=0,END ANG=360
ANGLE VEC=<1,0,0>
DIRECTION=CCW
SHOW FEATURE PARAMETERS=NO
SHOW CONTACT PARAMETERS=YES
NUMHITS=10,DEPTH=0.065,PITCH=0
SAMPLE METHOD=SAMPLE_HITS
SAMPLE HITS=0,SPACER=0
AVOIDANCE MOVE=BOTH,DISTANCE=1
FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
SHOW HITS=NO
                         (CODE REMOVED TO SAFE SPACE FOR FORUM)-WAS CODE FOR NON-IMPORTANT CIRCLES NOT AFFECTED
CIR20-NEXT ADJACENT CLEARANCE HOLE =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
THEO/<5.156,14.295,0>,<0,0,1>,0.56
ACTL/<5.1596,14.2993,-0.0119>,<0,0,1>,0.5678
TARG/<5.156,14.295,0>,<0,0,1>
START ANG=0,END ANG=360
ANGLE VEC=<1,0,0>
DIRECTION=CCW
SHOW FEATURE PARAMETERS=NO
SHOW CONTACT PARAMETERS=YES
NUMHITS=10,DEPTH=0.065,PITCH=0
SAMPLE METHOD=SAMPLE_HITS
SAMPLE HITS=0,SPACER=0
AVOIDANCE MOVE=BOTH,DISTANCE=1
FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
SHOW HITS=NO
CIR21-TOP LEFT CLEARANCE HOLE =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
THEO/<5.276,15.235,-0.315>,<0,0,1>,0.56
ACTL/<5.28,15.239,-0.3269>,<0,0,1>,0.566
TARG/<5.276,15.235,-0.315>,<0,0,1>
START ANG=0,END ANG=360
ANGLE VEC=<1,0,0>
DIRECTION=CCW
SHOW FEATURE PARAMETERS=NO
SHOW CONTACT PARAMETERS=YES
NUMHITS=10,DEPTH=0.065,PITCH=0
SAMPLE METHOD=SAMPLE_HITS
SAMPLE HITS=0,SPACER=0
AVOIDANCE MOVE=BOTH,DISTANCE=1
FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
SHOW HITS=NO
CIR22-TOP RIGHT CLEARANCE HOLE =FEAT/CONTACT/CIRCLE/DEFAULT,CARTESIAN,IN,LEAST_SQR
THEO/<6.276,15.235,-0.2563>,<0,0,1>,0.56
ACTL/<6.2796,15.2389,-0.2681>,<0,0,1>,0.5663
TARG/<6.276,15.235,-0.2563>,<0,0,1>
START ANG=0,END ANG=360
ANGLE VEC=<1,0,0>
DIRECTION=CCW
SHOW FEATURE PARAMETERS=NO
SHOW CONTACT PARAMETERS=YES
NUMHITS=10,DEPTH=0.065,PITCH=0
SAMPLE METHOD=SAMPLE_HITS
SAMPLE HITS=0,SPACER=0
AVOIDANCE MOVE=BOTH,DISTANCE=1
FIND HOLE=DISABLED,ONERROR=NO,READ POS=NO
SHOW HITS=NO
MOVE/POINT,NORMAL,<5.5233,2.3556,4.4173>
MOVE/INCREMENT,<0.0007,-0.0011,6>
MOVE/INCREMENT,<9,0.0005,-0.0011>
MOVE/POINT,NORMAL,<17.1669,-20.2313,10.1206>
LOADPROBE/ 1MM X 40MM
TIP/T1A0B0, SHANKIJK=0.0001, -0.0002, 1, ANGLE=0.003
CALIBRATE ACTIVE TIP WITH FEAT_ID=
SPH1-CALIBRATION TOOL =FEAT/SPHERE,CARTESIAN,OUT
THEO/<2.6061,-8.3221,4.4226>,<0,0,1>,0.9813
ACTL/<2.5995,-8.3296,4.4315>,<-0.0000314,0.0007901,0.9999997>,0.9841
MEAS/SPHERE,13
HIT/BASIC,NORMAL,<2.6428,-8.3562,4.9107>,<0.0749079,-0.0696,0.9947586>,<2.6318,-8.3602,4.9217>,USE THEO=YES
MOVE/CIRCULAR
HIT/BASIC,NORMAL,<2.8644,-8.5553,4.769>,<0.5260029,-0.4749804,0.7054889>,<2.8555,-8.5614,4.7817>,USE THEO=YES
MOVE/CIRCULAR
HIT/BASIC,NORMAL,<2.9066,-8.0994,4.7401>,<0.6124523,0.453895,0.6472105>,<2.8 97,-8.1022,4.7508>,USE THEO=YES
MOVE/CIRCULAR
              (CODE REMOVED TO SAFE SPACE FOR FORUM---WAS REST OF SPHERE-CAL CODING)
ENDMEAS/
​
  • Okay....if this helps anybody help me solve this problem, I might be onto something? So, I had to draw 2 different cal-spheres in the program to get each probe change to work properly. I don't know why, but if I tried using the same sphere for both single tip calibrations, it just wouldn't work. So I drew 2 different cal spheres when I first made this program a few weeks ago. But its confusing me because when I open the second single tip calibration command in the program it's using the first sphere for both calibrations, not 1 each. So it might be getting screwed up somehow that way...
  • When you created a second qual tool, did you make anything about it different other than name? Is it different in size, length, holding direction from your original? Did you use the same 1 tip to identify the location of each qual sphere?
    Additionally, what probing system are you using, it sounds like you are manually building/breaking them down in production?.
    Just a note: don't use any special characters in naming convention other than _ (underscore), don't even use spaces. It will help prevent odd occurrences.
  • @R2ah1ze1I I'm not exactly sure what I did (I made this program a few weeks ago and am just now adapting it for the loops). I know that I copied the sphere...and then just pasted it for the second calibration. I'm thinking that my problem should have been probing it the second time with the second probe. (I prob created the 2nd sphere or did the second autocal box with the first probe still selected). I am NOT breaking the tools down at all, but I am manually swapping them out (I don't have a tool changer). Even if I did have a tool changer, I was under the impression that I would still have to do an autocal or a single tip calibration for whatever probe I just changed, correct? I'm pretty sure I copied and pasted the sphere and didn't change the name at all.
  • , I definitely just checked and verified that there are 2 different spheres used for the single tip calibration. Before you commented I actually had it slightly different? There were 2 spheres, and it looked like one was on top of each other. I must have measured the spheres with a wrong tip in the wrong place. (probably measured with the 40mm long probe when the comp thought it had the 20mm long probe in)...anyways, I deleted both spheres and both single tip calibration commands. I then re-probed a sphere with the appropriate tool in the code, and they matched up now! There is sphere 1, and sphere 2. They are in the same place (because that's where the cal-sphere is obv). Sphere 1 is used for the first tool change, and sphere 2 is used for the second tool change. The problem still occurs though! It will inspect the first part with the first tool, change probes to the second, longer probe, recal the second tool, probe the first part, move over to the operator so I can change to the original 4mmx20mm probe. It then recals the first probe, and then does the measuring on the second part, then it pulls up in Z again, moves over and to me so that I can change it once again to the longer second probe. At the end of the program it pulls up in Z, does another move point to me so I can rechange the probe yet to the original first one. Then the loop and program stops.
  • Okay, , I did check, and yes, the two spheres are DIFFERENT in both size and location!!! It's not off by much, just a lil, but it's there, position is worse than diameter, off maybe a few thousandths, idk. I'm confused how this is even happening because I actually probed the ball in both cases to make the spheres??? Is the proper procedure to probe the spheres? Or am I supposed to plug in the machine coordinates and actual size of the ball into the program as a constructed sphere?? Or...another option would be to probe the sphere, but just alter the coordinates and size of the spheres nominal values after hitting done? The sphere being off even though I probed it slowly and IMO in a decent manner kind of freaks me out...makes me wonder if I've been throwing off bad part reports for the CMM. *scared*
  • Are you using TP20 probing? If you have magnetic coupling, you should be able to calibrate 1 time prior to running and then switch between your probe tips by manually swapping the magnetic components. How are you locating your calibration sphere? It should be performed in a repeatable fashion, same tip/angle/settings to help reduce error. (Good topic to search up is 'Master Probe') Is your calibration sphere attached to your fixture, or is it in a dedicated area on your CMM surface plate?
  • Yep! I do have magnetic coupling, but it's not a TP20, it's the TESA version, but almost identical. Still, I was told by multiple people including operators and people from Hexagon that a calibration is required every time a tool is swapped, even if it is magnetic. I think the problem I'm having is coming from the "calibrate single probe tip". I changed it to autocal in my program and am running it right now. I've looked into making a "Master/Main Probe"...the problem with that right now is that I only have 2 probe modules and no probe rack. I'm begging management for more probe modules basically.
  • no rack MAY be why they told you to calibrate after swapping.
  • OKAY!!! SO...I've *fixed* the problem. I say *fixed* because I didn't solve it using the exact parameters that I started out the problem with... I basically decided to just ditch "Single Probetip Calibrate" and instead inserted an auto-cal with my custom parameter sets (mainly just calibrating 0,0 only and the hit numbers and small things like that, the probe had like 7 angles originally) This was NOT an issue with looping, where my loop was, or anything to do with that. After I changed from single probetip cal to auto-cal everything lined up perfectly. There were a couple of variables that likely contributed to the original problems failure. First, the 2 spheres I drew for the single tip calibrate were NOT the same size and location. They were close, but by no means perfect. In fact, during the first tool change and calibrate, whenever the new tool would come down to make contact with the probe for the first touch, it wasn't even in the middle of the sphere, not even close!!! Also, during the second probe change, every now and then it would contact the probe too forcefully/ when it was done probing and trying to pull up it would just soft crash into the sphere. I am NOT the biggest fan of single probetip calibration now...I have it in a few other programs without looping, and I'm def going to go back for a closer look at those programs to make sure everything is working right, or I am just going to change it outright to avoid any future un-pleasantries.