I would also clean the ball stylus (once, the ball was unglued, but it was not visible- it fall down during the cleaning).
I would also put a little oil on little spheres on the "cup" of the LSPX.
I would try the block test with A0B90.
Check for cracks on the probe module. Also check the magnet for cracks or metal shavings. Maybe even try switching the suspect one with one that calibrated good.
I would measure the gage block along X with A0B0, and along Y with A0B90 (along Y with A0B0 is already done).
If the defect turns with the probe, it comes from the probe.
If the defect still the same after the head rotation, the problem comes from the cmm (dirts in air bearings or on pathes for example - it can be serious !!!!!)