High STD in 1 tip during calibration

Hello all.

I am getting 0.009+ mm of std while calibrating a tip in my global DEA. This is the case with only this tip/probe:



The 0.5mm balls are giving good results while the 2x40mm styli not.

The setup is as follows:



TP200 Renishaw head with a standard module.

I have tried more thna one styli / module / star config... Still not good results.

Based on REnishaw's instructions I don't see the styli being too long or too big for it to work...​

Any ideas here guys? Thanks in advance for any input
Parents
  • Hi ,
    I would say the horizontal length is to long for a TP200 SF.
    From renishaw :
    {"data-align":"none","data-size":"large","data-attachmentid":535696}


    This is nice info thanks Slight smile For some reason it is tough to find this lol:

    EO module also does not do the job... so effectively a 40mm styli cannot be used in star shape on tp200 right?
Reply
  • Hi ,
    I would say the horizontal length is to long for a TP200 SF.
    From renishaw :
    {"data-align":"none","data-size":"large","data-attachmentid":535696}


    This is nice info thanks Slight smile For some reason it is tough to find this lol:

    EO module also does not do the job... so effectively a 40mm styli cannot be used in star shape on tp200 right?