I am using a TP-20. I got a new tip size we havent used before (.7mm x 20mm ). I built a new probe file it is correct except HP-TB-2 is not in the list anymore, so i am using a HP-TBe. I have done this with other tip builds without a problem. When i try to calibrate all the tip angles over half come up with the error probe diameter exceeds limits, i try re-running those angles but get same results any suggestions?
tiny tip, SLOW DOWN the touch speed, 20mm long, gonna flex, which will give you smaller diameter results. Also, you need the lightest tension TP20, a standard module isn't much good for that small of a tip.
What is the size tolerance you are using for acceptable results?
tiny tip, SLOW DOWN the touch speed, 20mm long, gonna flex, which will give you smaller diameter results. Also, you need the lightest tension TP20, a standard module isn't much good for that small of a tip.
What is the size tolerance you are using for acceptable results?