I am using a TP-20. I got a new tip size we havent used before (.7mm x 20mm ). I built a new probe file it is correct except HP-TB-2 is not in the list anymore, so i am using a HP-TBe. I have done this with other tip builds without a problem. When i try to calibrate all the tip angles over half come up with the error probe diameter exceeds limits, i try re-running those angles but get same results any suggestions?
Any tip we have smaller than 1mm I start at 10 degrees when calibrating. But all of our tips that are that small are also short, which causes the shank to hit the cal. sphere before the ruby does when trying to start at zero.
Any tip we have smaller than 1mm I start at 10 degrees when calibrating. But all of our tips that are that small are also short, which causes the shank to hit the cal. sphere before the ruby does when trying to start at zero.