Ok two things:
1: I am pretty sure this isn't your problem, but make 100% certain you have a probe build that doesn't contain a SPACE at the end of the probe name.
2: PDCMIS thinks it's smarter than you are. I had this happen with trying to use TP200 modules on a MCR20 (not possible).
--Hexagon has some backend validation that makes certain that you can't use a TP200 on a TP20 rack, even though it's physically possible.
--I noticed your probe #6 in the picture is the only probe using a TESASTAR branded module.
--Try rebuilding that probe, reference a TP20 module in the build instead of the TESASTAR.
It will likely work then.
PCDMIS is probably programmed to prohibit any probe BUT TP20's from being mounted on a MCR20 rack.