hexagon logo

Runout of Constructed Circles

Hello Great Gurus (or I guess people with many points, so lame!) 

I am trying to measure the runout of a constructed circle, seemingly the Demon things I should not be doing that as its not an option on the list of features.

I am currently creating a circle at the intersection between a constructed cone and a constructed plane.

 

Should be right there at the bottom of the list. All of the other constructed circles are.
But to be fair, I think the last time I did this was with 2016, which would let you measure almost anything. 

Any ideas?

  • It won't let you use a constructed intersection circle because that kind of feature doesn't have any hits - it is obtained mathematically and has perfect form.  Runout looks at all of the hits from the feature and determines if they are all within the specified tolerance boundary.

  • Any idea for a clever work around? 
    Technically there is a ring of hits "near" where they want the runout taken, but for this part they really do want the runout of where the cone and plane intersect.
    Maybe I can cheat and work up some pierce points from what I have. Sadly I have a feeling this part is VERY not round so It wont really be using most of the data, reducing my large dataset to just a few points. 

  • Do you have an analogue scanning probe or just touch-trigger?

  • Level and zero on your axis of rotation.  Report the polar radius of the pierce points.  Subtract the largest polar radius from your smallest polar radius to get runout.   To do this automatically will take some clever programming to keep track of the high and low points and to do the subtraction for you.  

  • We have have a SP25. However due to some internal managerial type decisions we are not using it for any scanning at this time Neutral face SP25 doing TTP hits is my life right now. 

    The current approach is using using 4 groups of 20 hits each to construct each level of the cone. This both allows us to avoid the tooling and "not go crazy with the program" 

  • That's an interesting idea. I think it would work, but not sure it would gain enough of an accuracy boost to put the time into the programming. 

  • You can use the constructed adjust filter to "move" the points onto section.  The following code is an example of how to do it.  I have 4 levels (20 hits per level) on a 30° incl cone.  I build a constructed feature set from the third level (hits 41-60) and then apply an adjust filter to "shift" it to a cross-section height of Z=20.  I then construct a circle from the adjust filter which I can then dimeson for runout.  (Note: I left out some of the hits to save space)...

    CYL1       =FEAT/CONTACT/CYLINDER/DEFAULT,CARTESIAN,IN,LEAST_SQR
                THEO/<0,0,30>,<0,0,1>,6,10
                ACTL/<0,0,30>,<0,0,1>,6,10
                TARG/<0,0,30>,<0,0,1>
                START ANG=0,END ANG=360
                ANGLE VEC=<1,0,0>
                DIRECTION=CCW
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
                DATDEF/A,FEATURES=CYL1,,
    PNT0       =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-9.464,0,12>,<-0.9659258,0,0.258819>
                ACTL/<-9.464,0,12>,<-0.9659258,0,0.258819>
                TARG/<-9.464,0,12>,<-0.9659258,0,0.258819>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT1       =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-9.001,2.925,12>,<-0.9186502,0.2984871,0.2588191>
                ACTL/<-9.001,2.925,12>,<-0.9186502,0.2984871,0.2588191>
                TARG/<-9.001,2.925,12>,<-0.9186502,0.2984871,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT2       =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-7.657,5.563,12>,<-0.7814502,0.5677572,0.2588191>
                ACTL/<-7.657,5.563,12>,<-0.7814502,0.5677572,0.2588191>
                TARG/<-7.657,5.563,12>,<-0.7814502,0.5677572,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    
    ....
    
    PNT17      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-5.563,-7.657,12>,<-0.5677572,-0.7814502,0.2588191>
                ACTL/<-5.563,-7.657,12>,<-0.5677572,-0.7814502,0.2588191>
                TARG/<-5.563,-7.657,12>,<-0.5677572,-0.7814502,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT18      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-7.657,-5.563,12>,<-0.7814502,-0.5677572,0.2588191>
                ACTL/<-7.657,-5.563,12>,<-0.7814502,-0.5677572,0.2588191>
                TARG/<-7.657,-5.563,12>,<-0.7814502,-0.5677572,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT19      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-9.001,-2.925,12>,<-0.9186502,-0.2984871,0.2588191>
                ACTL/<-9.001,-2.925,12>,<-0.9186502,-0.2984871,0.2588191>
                TARG/<-9.001,-2.925,12>,<-0.9186502,-0.2984871,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT20      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-7.642,-2.483,17.333>,<-0.9186502,-0.2984871,0.2588191>
                ACTL/<-7.642,-2.483,17.333>,<-0.9186502,-0.2984871,0.2588191>
                TARG/<-7.642,-2.483,17.333>,<-0.9186502,-0.2984871,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT21      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-6.5,-4.723,17.333>,<-0.7814502,-0.5677572,0.2588191>
                ACTL/<-6.5,-4.723,17.333>,<-0.7814502,-0.5677572,0.2588191>
                TARG/<-6.5,-4.723,17.333>,<-0.7814502,-0.5677572,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT22      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-4.723,-6.5,17.333>,<-0.5677572,-0.7814502,0.2588191>
                ACTL/<-4.723,-6.5,17.333>,<-0.5677572,-0.7814502,0.2588191>
                TARG/<-4.723,-6.5,17.333>,<-0.5677572,-0.7814502,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT23      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-2.483,-7.642,17.333>,<-0.2984871,-0.9186502,0.2588191>
                ACTL/<-2.483,-7.642,17.333>,<-0.2984871,-0.9186502,0.2588191>
                TARG/<-2.483,-7.642,17.333>,<-0.2984871,-0.9186502,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    
    ...
    
    PNT37      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-6.5,4.723,17.333>,<-0.7814502,0.5677572,0.2588191>
                ACTL/<-6.5,4.723,17.333>,<-0.7814502,0.5677572,0.2588191>
                TARG/<-6.5,4.723,17.333>,<-0.7814502,0.5677572,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT38      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-7.642,2.483,17.333>,<-0.9186502,0.2984871,0.2588191>
                ACTL/<-7.642,2.483,17.333>,<-0.9186502,0.2984871,0.2588191>
                TARG/<-7.642,2.483,17.333>,<-0.9186502,0.2984871,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT39      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-8.035,0,17.333>,<-0.9659258,0,0.258819>
                ACTL/<-8.035,0,17.333>,<-0.9659258,0,0.258819>
                TARG/<-8.035,0,17.333>,<-0.9659258,0,0.258819>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT40      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-6.606,0,22.667>,<-0.9659258,0,0.258819>
                ACTL/<-6.606,0,22.667>,<-0.9659258,0,0.258819>
                TARG/<-6.606,0,22.667>,<-0.9659258,0,0.258819>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT41      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-6.283,2.041,22.667>,<-0.9186502,0.2984871,0.2588191>
                ACTL/<-6.283,2.041,22.667>,<-0.9186502,0.2984871,0.2588191>
                TARG/<-6.283,2.041,22.667>,<-0.9186502,0.2984871,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT42      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-5.344,3.883,22.667>,<-0.7814502,0.5677572,0.2588191>
                ACTL/<-5.344,3.883,22.667>,<-0.7814502,0.5677572,0.2588191>
                TARG/<-5.344,3.883,22.667>,<-0.7814502,0.5677572,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT43      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-3.883,5.344,22.667>,<-0.5677572,0.7814502,0.2588191>
                ACTL/<-3.883,5.344,22.667>,<-0.5677572,0.7814502,0.2588191>
                TARG/<-3.883,5.344,22.667>,<-0.5677572,0.7814502,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT44      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-2.041,6.283,22.667>,<-0.2984871,0.9186502,0.2588191>
                ACTL/<-2.041,6.283,22.667>,<-0.2984871,0.9186502,0.2588191>
                TARG/<-2.041,6.283,22.667>,<-0.2984871,0.9186502,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT45      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<0,6.606,22.667>,<0,0.9659258,0.258819>
                ACTL/<0,6.606,22.667>,<0,0.9659258,0.258819>
                TARG/<0,6.606,22.667>,<0,0.9659258,0.258819>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT46      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<2.041,6.283,22.667>,<0.2984871,0.9186502,0.2588191>
                ACTL/<2.041,6.283,22.667>,<0.2984871,0.9186502,0.2588191>
                TARG/<2.041,6.283,22.667>,<0.2984871,0.9186502,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT47      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<3.883,5.344,22.667>,<0.5677572,0.7814502,0.2588191>
                ACTL/<3.883,5.344,22.667>,<0.5677572,0.7814502,0.2588191>
                TARG/<3.883,5.344,22.667>,<0.5677572,0.7814502,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT48      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<5.344,3.883,22.667>,<0.7814502,0.5677572,0.2588191>
                ACTL/<5.344,3.883,22.667>,<0.7814502,0.5677572,0.2588191>
                TARG/<5.344,3.883,22.667>,<0.7814502,0.5677572,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT49      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<6.283,2.041,22.667>,<0.9186502,0.2984871,0.2588191>
                ACTL/<6.283,2.041,22.667>,<0.9186502,0.2984871,0.2588191>
                TARG/<6.283,2.041,22.667>,<0.9186502,0.2984871,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT50      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<6.606,0,22.667>,<0.9659258,0,0.258819>
                ACTL/<6.606,0,22.667>,<0.9659258,0,0.258819>
                TARG/<6.606,0,22.667>,<0.9659258,0,0.258819>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT51      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<6.283,-2.041,22.667>,<0.9186502,-0.2984871,0.2588191>
                ACTL/<6.283,-2.041,22.667>,<0.9186502,-0.2984871,0.2588191>
                TARG/<6.283,-2.041,22.667>,<0.9186502,-0.2984871,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT52      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<5.344,-3.883,22.667>,<0.7814502,-0.5677572,0.2588191>
                ACTL/<5.344,-3.883,22.667>,<0.7814502,-0.5677572,0.2588191>
                TARG/<5.344,-3.883,22.667>,<0.7814502,-0.5677572,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT53      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<3.883,-5.344,22.667>,<0.5677572,-0.7814502,0.2588191>
                ACTL/<3.883,-5.344,22.667>,<0.5677572,-0.7814502,0.2588191>
                TARG/<3.883,-5.344,22.667>,<0.5677572,-0.7814502,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT54      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<2.041,-6.283,22.667>,<0.2984871,-0.9186502,0.2588191>
                ACTL/<2.041,-6.283,22.667>,<0.2984871,-0.9186502,0.2588191>
                TARG/<2.041,-6.283,22.667>,<0.2984871,-0.9186502,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT55      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<0,-6.606,22.667>,<0,-0.9659258,0.258819>
                ACTL/<0,-6.606,22.667>,<0,-0.9659258,0.258819>
                TARG/<0,-6.606,22.667>,<0,-0.9659258,0.258819>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT56      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-2.041,-6.283,22.667>,<-0.2984871,-0.9186502,0.2588191>
                ACTL/<-2.041,-6.283,22.667>,<-0.2984871,-0.9186502,0.2588191>
                TARG/<-2.041,-6.283,22.667>,<-0.2984871,-0.9186502,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT57      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-3.883,-5.344,22.667>,<-0.5677572,-0.7814502,0.2588191>
                ACTL/<-3.883,-5.344,22.667>,<-0.5677572,-0.7814502,0.2588191>
                TARG/<-3.883,-5.344,22.667>,<-0.5677572,-0.7814502,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT58      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-5.344,-3.883,22.667>,<-0.7814502,-0.5677572,0.2588191>
                ACTL/<-5.344,-3.883,22.667>,<-0.7814502,-0.5677572,0.2588191>
                TARG/<-5.344,-3.883,22.667>,<-0.7814502,-0.5677572,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT59      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-6.283,-2.041,22.667>,<-0.9186502,-0.2984871,0.2588191>
                ACTL/<-6.283,-2.041,22.667>,<-0.9186502,-0.2984871,0.2588191>
                TARG/<-6.283,-2.041,22.667>,<-0.9186502,-0.2984871,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT60      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-4.924,-1.6,28>,<-0.9186502,-0.2984871,0.2588191>
                ACTL/<-4.924,-1.6,28>,<-0.9186502,-0.2984871,0.2588191>
                TARG/<-4.924,-1.6,28>,<-0.9186502,-0.2984871,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT61      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-4.188,-3.043,28>,<-0.7814502,-0.5677572,0.2588191>
                ACTL/<-4.188,-3.043,28>,<-0.7814502,-0.5677572,0.2588191>
                TARG/<-4.188,-3.043,28>,<-0.7814502,-0.5677572,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT62      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-3.043,-4.188,28>,<-0.5677572,-0.7814502,0.2588191>
                ACTL/<-3.043,-4.188,28>,<-0.5677572,-0.7814502,0.2588191>
                TARG/<-3.043,-4.188,28>,<-0.5677572,-0.7814502,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    
    ...
    
    PNT77      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-4.188,3.043,28>,<-0.7814502,0.5677572,0.2588191>
                ACTL/<-4.188,3.043,28>,<-0.7814502,0.5677572,0.2588191>
                TARG/<-4.188,3.043,28>,<-0.7814502,0.5677572,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT78      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-4.924,1.6,28>,<-0.9186502,0.2984871,0.2588191>
                ACTL/<-4.924,1.6,28>,<-0.9186502,0.2984871,0.2588191>
                TARG/<-4.924,1.6,28>,<-0.9186502,0.2984871,0.2588191>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    PNT79      =FEAT/CONTACT/VECTOR POINT/DEFAULT,CARTESIAN
                THEO/<-5.177,0,28>,<-0.9659258,0,0.258819>
                ACTL/<-5.177,0,28>,<-0.9659258,0,0.258819>
                TARG/<-5.177,0,28>,<-0.9659258,0,0.258819>
                SHOW FEATURE PARAMETERS=NO
                SHOW CONTACT PARAMETERS=NO
    CON1       =FEAT/CONE,CARTESIAN,OUT,LENG,NO
                THEO/<0,0,47.321>,<0,0,-1>,16,10.354,18.928
                ACTL/<0,0,47.321>,<0,0,-1>,16,10.354,18.928
                CONSTR/CONE,BFRE,PNT0,PNT79,PNT78,PNT77,PNT76,PNT75,PNT74,PNT73,PNT72,PNT71,PNT70,PNT69,PNT68,PNT67,PNT66,PNT65,PNT64,PNT63,PNT62,PNT61,PNT60,PNT59,PNT58,PNT57,PNT56,PNT55,PNT54,PNT53,PNT52,PNT51,PNT50,PNT49,PNT48,PNT47,PNT46,PNT45,PNT44,
                                 PNT43,PNT42,PNT41,PNT40,PNT39,PNT38,PNT37,PNT36,PNT35,PNT34,PNT33,PNT32,PNT31,PNT30,PNT29,PNT28,PNT27,PNT26,PNT25,PNT24,PNT23,PNT22,PNT21,PNT20,PNT19,PNT18,PNT17,PNT16,PNT15,PNT14,PNT13,PNT12,PNT11,PNT10,PNT9,PNT8,PNT7,$
                                 PNT6,PNT5,PNT4,PNT3,PNT2,PNT1,,
    SCN1       =FEAT/SET,CARTESIAN
                THEO/<0.084,-0.08,22.933>,<0,0,1>
                ACTL/<0.084,-0.08,22.933>,<0,0,1>
                CONSTR/SET,BASIC,PNT41,PNT60,PNT59,PNT58,PNT57,PNT56,PNT55,PNT54,PNT53,PNT52,PNT51,PNT50,PNT49,PNT48,PNT47,PNT46,PNT45,PNT44,PNT43,PNT42,,
    SCN2       =FEAT/SET,CARTESIAN
                THEO/<0.084,-0.08,22.933>,<0,0,1>
                ACTL/<0.018,-0.113,20>,<0,0,1>
                CONSTR/SET,ADJUSTFILTER,CONE,RADIAL,30,SCN1,CON1,SHOW HITS=NO
                FEATURE/0,0,47.321,<0,0,-1>
                PLANE/0,0,20,<0,0,1>
    CIR1       =FEAT/CIRCLE,CARTESIAN,OUT,LEAST_SQR,YES
                THEO/<0,0,20>,<0,0,1>,13.071
                ACTL/<0,0,20>,<0,0,1>,14.641
                CONSTR/CIRCLE,BF,2D,SCN2,,
                OUTLIER_REMOVAL/OFF,3
                FILTER/OFF,UPR=0
    FCFRNOUT1  =GEOMETRIC_TOLERANCE/STANDARD=ASME Y14.5,SHOWEXPANDED=YES,
                DESCRIPTION=ON,,
                DATUM_MATH=DEFAULT,
                UNITS=MM,OUTPUT=BOTH,ARROWDENSITY=100,
                SEGMENT_1,CIRCULAR_RUNOUT,0.01,A,<dat>,
                TEXT=OFF,CADGRAPH=ON,REPORTGRAPH=OFF,MULT=100,
                  NOMINAL:
                    HALF ANGLE=15,
                  MEASURED:
                    CIR1:0.000,
                DATUMS/REPORTDATUMSIZE=ON,
                  A(CYL1):NOM=6,+Tol=0.01,-Tol=0.01,
                FEATURES/CIR1,,

  • Oh that looks pretty good  
    I will give that a shot and report back. 

  • Keep in mind, the more points you can take and the closer you can measure your points to the real section height, the more accurate it will be.

  • A lot of CMM software will not allow form or eccentricity error doing this directly.  Do this:  measured line (autofeature) on cone, measured line (autofeature) on plane, intersect point.  Paste that with a pattern around axis.  Now with all those intersect points, you can get everything you want, at least in the less stupid versions of PC-DMIS