I am using a 6.35mm (.250") Probe to calibrate a simple 12" block
I keep getting a 12.500 reading.
What is the Edit Probe Data suppose to be in
X, Y, Z Center
Diameter & Thickness?
or does anyone know what could be the problem. I calibrated to the 1" ball w/in .001.
If probe comp. isn't turned on and your probe is exactly .25" (6.35mm) and you took points on both ends, you would see +/- 12.5". Is your probe comp turned on? Oh and I haven't touched a Faro arm in at least 10 years so anything I tell you may be totally useless.
If probe comp. isn't turned on and your probe is exactly .25" (6.35mm) and you took points on both ends, you would see +/- 12.5". Is your probe comp turned on? Oh and I haven't touched a Faro arm in at least 10 years so anything I tell you may be totally useless.