does anyone know? when you calibrate the star probe for the cmm-ve angles do you have to everytime go to cmm-ve and add angles? or just the first time?
You have to do it each time, and moreover, you have to do it each time after had answer "yes, the sphere move"...
Personnaly, I calibrate all angles between A0B90 and A90B90, and I use a rotary table to use other angles !
I have also a dull sphere that I use to check the calibration, with a auto focus point on the radius, with tip vector, and a circle on equator. I would like to enter automatically deviation values in the probe calibration, but I don't know how to do it... I have usually some µm (between 5 and 30 !) between contact calibration and vision check...
are you using a 32 bit or a 64 bit? i have a older machine with a 32 bit and it works fine but the new install came with a 64 bit and it takes 20 to 30 seconds to measure a line.
all documentation say not to use a 64 bit and i keep telling Hexagon and they say it should work, hexagon has been trying to get it to work now for 3 months now and i am very frustrated.
I work in 32bit, but our local tech (Hexagon) works on 64bit. I think that there's only one video card that works with cmmv and 64bit, (IDS Falcon, bt I'm not sure)