does anyone know? when you calibrate the star probe for the cmm-ve angles do you have to everytime go to cmm-ve and add angles? or just the first time?
You have to do it each time, and moreover, you have to do it each time after had answer "yes, the sphere move"...
Personnaly, I calibrate all angles between A0B90 and A90B90, and I use a rotary table to use other angles !
I have also a dull sphere that I use to check the calibration, with a auto focus point on the radius, with tip vector, and a circle on equator. I would like to enter automatically deviation values in the probe calibration, but I don't know how to do it... I have usually some µm (between 5 and 30 !) between contact calibration and vision check...
You have to do it each time, and moreover, you have to do it each time after had answer "yes, the sphere move"...
Personnaly, I calibrate all angles between A0B90 and A90B90, and I use a rotary table to use other angles !
I have also a dull sphere that I use to check the calibration, with a auto focus point on the radius, with tip vector, and a circle on equator. I would like to enter automatically deviation values in the probe calibration, but I don't know how to do it... I have usually some µm (between 5 and 30 !) between contact calibration and vision check...