2019 R2 + CMMVe problems

I create a new topic because the previous title was SP1, so this one is only 2019 R2, and be able to have a long life Wink.

2019 R2 SP2 :
During the optical calibration, you have to center the camera on concentric circles (Hexagon glass artifact).
The focus is mesured, and at the end, the cmm doesn't move to the focus position : is it a new parameter ?

During the optical calbration, the cmm moves in x and y at a very very slow speed (about 0.25 mm / s), it looks like the focus speed.
Is there a setting to change it ?

That's all for about 1 hour of tests, the rotary turns Slight smile.

Happy week-end, all !
Parents
  • Hi Jeffman

    Regarding the Slow Move during the Optical Calibration, I believe you are are seeing an issue we recently identified and fixed. I believe this fix is scheduled for the next SP. If it is the same issue, the slow moves go away if the setting Move to Focus Position is enabled in F5 Settings. This setting was incorrectly affecting the behaviour of the calibration
Reply
  • Hi Jeffman

    Regarding the Slow Move during the Optical Calibration, I believe you are are seeing an issue we recently identified and fixed. I believe this fix is scheduled for the next SP. If it is the same issue, the slow moves go away if the setting Move to Focus Position is enabled in F5 Settings. This setting was incorrectly affecting the behaviour of the calibration