You must qualify the same stylus twice. You can use QualifyTool to calibrate the disc diameter & CALPLA to calibrate the bottom face.
The reason is, the deflection for the disc probe diameter is zero in Z & deflection for the disc probe face is zero in X & Y.
I've never done it that way, but I'm told it works.
You must qualify the same stylus twice. You can use QualifyTool to calibrate the disc diameter & CALPLA to calibrate the bottom face.
The reason is, the deflection for the disc probe diameter is zero in Z & deflection for the disc probe face is zero in X & Y.
I've never done it that way, but I'm told it works.