I've been using the CMM for a about 4 years now in a job shop. We only use it when we *need* to (when we need to produce an inspection report for the customer). I'm the only CMM operator. Running 2012 MR1 CAD. I have several parts that I need to inspect and report on and have the same issues with, but I'll direct this thread toward one in particular.
I need to measure true position of some threaded holes and straight holes that will have pins pressed in them. I'm using the CMM currently to do an in-process inspection so the machinist knows he can run the parts. These are fairly complex parts with multiple operations. The threaded holes (measured before threads) are okay because they have a tolerance of .2mm to ABC. The pin hole (for 6mm pin) positions look strange and are out of spec. [TP|⌀.08|A|B|C]
I'm pretty sure my alignment is good. I leveled, rotated and set XYZ to Datums A, B and C (all planes). My part is up on 3 points and against a stop (repeatable). I'm taking 6 hits on 3 circles per hole and using the circles to create cylinders, then taking TP on the cylinders. I calibrate the probe every time I change it (manual). (I'm switching between 2 jobs today that both require separate probes.) I made sure my air bearing ways were clean, along with the probe, calibration tool and table.
I can post code or portions of the print if needed, but I'm mostly wondering what the general causes of error are in a CMM program? I've searched through many forum posts to find out what I could be missing in my program. The machinists never like when I have to inspect something because they don't trust my numbers. I try to verify my numbers with the height gauge, but if the numbers aren't exact then everyone has a theory (or 3) as to why things didn't come out right. Often someone will pick whichever they *feel* is better and go with that method (CMM or height gage), even if the other shows numbers out of tolerance.
What else can I do to be confident in my program and my TP numbers?
Any help would be greatly appreciated.
When they dont trust the bad dimensions - then how can they trust the good ones ?
Are the cylinders perpendicular to the surface they intersect?
Using Legacy or XactMeasure for reporting? XM has additional check boxes that will affect the reporting.
When reporting is the Position reported using Worst, Average, or Start/End points? Each one can produce a different result.
Do the linear deviations "match up" with the reported Position deviation?
Are points taken at each level at the same height?
Are separate circles taken and cylinders constructed, or is auto-cylinder used?
What algorithm is used to create the cylinders?
DIM LOC1= POSITION OF CYLINDER PN1A USE AXIS=AVERAGE REF LENGTH=0.000 UNITS=MM ,$ GRAPH=OFF TEXT=OFF MULT=10.00 OUTPUT=BOTH FIT TO DATUMS=ON DEV PERPEN CENTERLINE=OFF DISPLAY=DIAMETER AX NOMINAL +TOL -TOL BONUS MEAS DEV OUTTOL X 18.050 18.095 0.045 Z -142.000 -141.983 0.017 DF 6.000 0.000 0.030 5.978 -0.022 0.000 -----#-------------- TP RFS 0.080 0.000 0.097 0.097 0.017 -------------------> END OF DIMENSION LOC1
Ideally, yes. The tapped holes are drilled. The pin holes were milled (interpolated), but now he's spot drilling, drilling and using a reamer.
So the holes are or are not perpendicular? Report each ones' Perp and find out.
I'm using Xact. What boxes are you referring to?
"Use Current Alignment" ON or OFF, "Perp to Centerline" ON or OFF. Each check box being ON/OFF can produce different results.
I noticed those on the report. I'm not sure how to change this so they are all measured correctly. Sometimes it uses the start point and sometimes the end points. I think it always does Worst.
Change this in the code - hover over "Worst", click when it becomes active, and the other selections show up. For me part function or purpose of measurement determines which I select. I do use "Average" more frequently.
I checked the math of the deviations vs the TP dev. Most of the numbers match up. The ones that are off look to be a rounding difference.
When I used the height gauge to verify I was off by about .03 in one direction.
0.03mm is A LOT when the tolerance is only 0.08mm. Do the linear measurements make sense?
Yes (see below)
The circles are measured at the same nominal depths and then a cylinder is constructed. In a previous program for this part I used auto-cyl.
I'm not sure what this is referring to.. In the cyl edit window it shows Best fit and LEAST_SQR
Search HELP (Best Fit) to answer this question. The closer the Form of the cylinder gets to being perfectly round the less difference there will be between each algorithm.