hi everyone.

  I have a problem during calibration time. I work with a CMM machine that has 3 probe slots. Durind calibration, the x,y,z  coordinates can deviate. I don't know actual problem. so, I'm unsure how to solve it , kindly, if anyone can provide a solution it would be help for me

Parents
  • I would say your set parameters is wrong (maybe "sphere moved is on "yes", so that the calibration position the sphere with actual calibration params of each probe, instead of calibrating the probes.)

    If you want to position the sphere, then create a parameter set with A0B0, to locate the sphere, then execute other tips with "No"

Reply
  • I would say your set parameters is wrong (maybe "sphere moved is on "yes", so that the calibration position the sphere with actual calibration params of each probe, instead of calibrating the probes.)

    If you want to position the sphere, then create a parameter set with A0B0, to locate the sphere, then execute other tips with "No"