Datapage+ Reporting opposite of PCDMIS

Does anyone know why PCDMIS would report a value as Inboard, but when I create my report in datapage it says Outboard? I'll post a pic when I get a minute, one of my engineers just noticed this. Very frustrating!!!Angry
Parents

  • You need to look at the T value in both examples. They both say the same thing (inbound value). If you then look at the -y value it is shorter than the nominal (inbound). The measured value -749.265 is closer to the origin than the nominal value of -750.348. The data is presented correctly.



    Attached Files
Reply

  • You need to look at the T value in both examples. They both say the same thing (inbound value). If you then look at the -y value it is shorter than the nominal (inbound). The measured value -749.265 is closer to the origin than the nominal value of -750.348. The data is presented correctly.



    Attached Files